TANGO Test Log


Company : THALES AIR DEFENCE
Project : TANGO Aquila Test Bench
Program : STAR_DIGITAL_GENERATION_TEST 47 068 402
Serial Number : No Serial Number
Start Time : 05/13/13   16:34:43

Task 2 : CKTST 2 : POWER SUPPLY TEST

# Name Pin Unit Min Max Result Status
001 test 2_1 : +5 V voltage short circuit P1-36/P1-37 Ohm 10.000 357.683 Pass
002 test 2_2 : +15 V voltage short circuit P1-40/P1-41 Ohm 10.000 348.374 Pass
003 test 2_3 : -5,2 V voltage short circuit P1-46/P1-47 Ohm 10.000 32458.070 Pass
# Name Pin Unit Tolerence Value Result Status
004 test 2_4 : +5 V current level P1-36/P1-37 Amp +/ -0.200 0.860 0.689 Pass
005 test 2_5 : +15 V current level P1-40/P1-41 Amp +/ -0.050 0.531 0.394 Fail*
006 test 2_6 : -5,2 V current level P1-46/P1-47 Amp +/ -0.050 0.153 0.165 Pass
007 test 2_7 : +5 V power supply voltage P1-36/P1-37 Volt +/ -0.250 5.000 5.240 Pass
008 test 2_8 : +15 V power supply voltage P1-40/P1-41 Volt +/ -0.750 15.000 15.191 Pass
009 test 2_9 : -5,2 V power supply voltage P1-46/P1-47 Volt +/ -0.250 -5.200 -5.185 Pass

Task 5 : CKTST 5 : EXTERNAL PHASE TEST


...Application aborted.



... PROGRAM ABORTED ...