Temperature Testing

Andrew Y.
Bristol, PA

Oct 8, 2015
5 Posts

0  |  0  

Re: Temperature Testing

I am trying to run 4 units in a over temperature and just can't seem to grasp the flow control.  I have the profile driver in the project which allows me to run a profile.  I created a project called temperature and added tests called cold, hot and room.  This allowed me, per unit, to add temperature sets to the test profile.

However I programmed myself into a corner, with 4 UUTs if I put any of those temperature tests into the profile it will execute the test-cold-test-hot-test-room for the UUT, then switch to the next UUT then having to re-soak the UUT at that Temp.  If I put any of the temperature stuff into the UUT project the same thing will happen.  

The desired result is:
Goto Room Temp
UUT1 Test, UUT2 Test, UUT3 Test, UUT4 Test
Goto Cold Temp
UUT1 Test, UUT2 Test, UUT3 Test, UUT4 Test
Goto Hot Temp
UUT1 Test, UUT2 Test, UUT3 Test, UUT4 Test
Goto Room Temp

I know it there has to be a simple solution involving the test executive driver, but just can't see it.  Any help would be great.

Thanks,
Andrew

Solution Available
Andrew Y.
Bristol, PA

Oct 8, 2015
5 Posts

1  |  0  

Re: Temperature Testing

Found it.

Need to set:
aUutSwitchLevelTask to 2.

Very clever.....